|
 |
 |
DEVELOPMENT AND
VALIDATION OF A CHROMATOGRAPHIC METHOD
|
(This manuscript is part of
the HPLC
Course given by Dr. Shulamit Levin, Medtechnica)
 |
-
REFERENCE STANDARDS:
-
1.a. Established source and
known grade (DMF
- Drug Master File
or COA - Certificate of Analysis)
- 1.b. Percentage of purity
from assay will
be taken into account in the calculations.
-
1.c. Amount of residual
compounds (GC, heavy
metals, inorganic salts, water,
residual solvents, weight loss)
should
be known.
-
CHOICE OF
STANDARDIZATION: EXTERNAL OR
INTERNAL
-
2.a. Simple formations and
sample preparation:
external
standard
-
2.b. Gas chromatography,
bio-studies or complex
medium and sample
preparation: internal
standard.
-
SELECTION AND DEVELOPMENT
OF THE CHROMATOGRAPHIC
METHOD:
-
3.a. The chromatographic
method from the literature
(Pharmacopoeia) must be the first
choice, with freedom of coloumn
choice
(according to the Lx tables in the USP).
Peak shape should be
examined in
this step already (see 4.a.).
- 3.b. The method should be
adjusted to be stability
indicating and selective, if this is
impossible, a new method should
be developed
and validated to be stability indicating.
- 3.c. Potential impurities
and degradation
products should be used with the sample medium.
-
3.d. Peak purity of the active
substance is
checked (by phodo-diode array detector to verify
that the method is selective,
and a single
component peak is quantified.
- 3.e. The analysis time range
should be kept
practical, no more than 40 min, i.e., the retention
parameter should be kept k(first)
>= 2 and k(last)
<= 40 (flow-rate,
column length and solvent
composition can be adjusted to
speed up
the separation).
- 3.f. Final method is defined
by the following
parameters:
-
Column type: geometry and
brand
-
Mobile phase
-
Flow rate and pressure
-
Temperature
-
Sample volume
-
Detection
-
Sample preparation(ID)
-
COLUMN AND INSTRUMENT
PERFORMANCE:
SYSTEM SUITABILITY TEST
The following parameters must be
checked,
preferably during method development:
-
4.a. Number of theoretical
plates, N>=2000.
[16(tR/W) ^2]
-
4.b. Tailing factor
0.5 <= T
=< 2.
If these requirements cannot be
fulfilled,
the column must be replaced. This test is done
during the method development
(3.a.).
- 4.c. Precision RSD
<= 1-2% in 10
injections of the standard. If RSD > 1%, the
autosampler should be checked,
or a leak
might be detected in the instrument.
- 4.d. Resolution, Rs
>= 2 between adjacent
peaks; [(tR2-tR1)/0.5(w1+w2)]
Note:
This step should be measured as
part of
adjustment of the method to be stability indicating and selective.
Sections 1 and 2 verify that the
chromatographic
system and column follow the requirements
of the analysis. The next steps
aim to
establish the method of analysis.
-
METHOD VALIDATION
-
5.a. Linearity: the
range of linearity
is checked by injections of 5-6 concentrations of the
reference standards (in
triplicates) below
and above the expected concentration of the
samples (20%-120%). The slope
and intercept
are determined. r >= 0.999, and the
range of linearity (in conc.
terms) is
established.
- 5.b. Accuracy is
measured as the (amount
measured)/(amount claimed) in the following
procedures:
-
In drug formulations: a
standard is added
to the placebo (80%, 100%, 120%), and the
recovery is measured
- Recovery from sample
preparation procedures
(such as dilution, extraction or
filtration) should be
established.
- Solution stability of the
samples (A fresh
sample is compared to one which was left in
solution for the entire
duration of the
test procedure).
-
5.c. Intermediate
precision or ruggedness:
5 measurements of the first step of accuracy under two
different occasions (two
columns, or two
operators).
- 5.d. Limit of detection
(LOD) and limit
of quantification (LOQ):
-
The intercept from the
linearity test (in
section 5.a) can be used for LOQ
measurements (intercept x 5 or
10).
- Analysis of repeatability
at the quantification
limit, using both sample and reference
standard. Concentrations that
yield 10-15%
RSD will be acceptable as LOQ concentrations.
- 5.e. Robustness of the
chromatographic
method: variables in this test can be the
following: column batch #,
solvent strength
in the mobile phase, temperature, flow rate,
pH of the mobile phase, ionic
strength
in the mobile phase, sample diluent,
injection volume, wavelength
of detection.
The parameter measured: response,
retention time, selectivity
and/or resolution.
NOTE:
If the step of method development is
done
systematically and documented well, robustness
can be established at early stages.
Robustness of the sample
preparation
procedure: variables in this test can be the
following: duration of extraction,
extraction
medium, filtration type, temperatures.
Parameter measured: accuracy.
Please note: This document is
the sole
property of Shula Levin. Reproduction of this document without the
express
permission of Shula Levin, for any purpose, is strictly forbidden

Home||
HPLC
Topics ||
LC-MS
TOPICS || HPLC
Course || Educational
|| Links
Courses
|| Personal
||
FTIR
||
Contact
Me
Copyright (C)2000 Dr. Shulamit
Levin
|