Based on the Quick Start Guide
34 Maple Street
Milford, MA 01757
WAT553-03TP, Revision 4

Put all inlets in the
appropriate solvents according to the rules of HPLC work (see guide for HPLC
work)
Power on the Separations
Module. Wait until STARTUP DIAGNOSTICS is over – do not open any door!
When the initial part of the
startup diagnostics routine is successful, the Main screen appears in
the front panel display. The startup diagnostics routine continues by
initializing the:
• Needle, syringe, and valves
• Carousel carrier
• Solvent management system
The status of these mechanical
diagnostics tests appear in the banner area of the Main screen, while
the diagnostics are running. When the startup diagnostics routine is
complete, the Separations Module enters the Idle
mode.
To perform a dry prime:
.Attach an empty syringe to the
prime/vent valve, then open the valve by turning it counterclockwise
1-2-turn.
2. Withdraw the
syringe plunger to pull solvent through the tubing. You may need to
exert force to pull the air and solvent through the system. Continue
until you pull all the air through the solvent line into the syringe.
3. Repeat steps 1 and 2 for
each solvent line you want to prime, then close the prime/vent valve.
Optional:
In the Enter a Duration field,
enter the length of time (in minutes) to prime the solvent management
system. Start with a value of 5 minutes. Press Continue.
When the system is dry, Waters
recommends that you prime using the solvent with the lowest viscosity
to help purge air from the lines, especially if the in-line vacuum
degasser is installed.
Perform a wet prime when you want
to change reservoirs or solvents, or if the Separations Module has been
idle for some time. Use an intermediate solvent such as water when you
change from buffers to high-organic-content solvents.
1. In the Status screen
Composition field, enter 100% for
the each solvent you want to use for wet priming..
2. Press the Direct Function screen key
in the Status screen. The Direct Functions menu appears.
3. Select the Wet Prime option, then
press Enter. The Wet
Prime dialog box
appears.
To control the operation of the
in-line vacuum degasser from the Status screen:
1. Press the Menu/Status key.
2. Select the Degasser mode field, then
press Enter to
display a list of degasser modes. The modes are:
2790:
•Normal– The degasser cycles on and off on a
pressure-regulated cycle
•Continuous– The degasser is always on
•Disable– The degasser is always off
2795:
On - The degasser is always on
Off- The degasser is off
4. Select the desired operating
mode, then press Enter.
5. Wet prime again
after vacuum pressure was established.
For reversed-phase HPLC
applications, use an aqueous plunger seal-wash solution with enough
organic content to inhibit bacterial growth. For example, use an 90%
water:10% methanol solution or an 90% water:10% acetonitrile
solution, depending on your application. For normal phase separations,
use a 50% methanol:50% IPA seal wash solution. If required, use an
intermediate solution to prevent immiscibility or precipitate problems
when you switch between normal phase and reversed-phase analytical
solvents.
To prime the plunger seal wash
pump:
1. Press the Diag screen key in the Main
screen. The Diagnostics screen appears.
2. Press the Prime SealWsh screen key,
then press the Start screen
key. The seal wash procedure begins. Push on a syringe plunger to force
seal wash solvent through the system if needed.
3. Press Close to return to the
Diagnostics screen.
The needle-wash pump flushes the
needle in the sample management system, preventing carryover of sample
between injections. The needle wash also extends the life of the
injector seals by removing buffered mobile phase and sample from the
needle..
To prime the needle-wash pump:
1. Press the Diag screen key in the Main
screen. The Diagnostics screen appears.
2. Press the Prime NdlWsh screen key. The
30-second needle-wash procedure begins. If solvent does not flow out of
the waste tube within 30 seconds, press the Start Again screen key.
Use a needle-wash solvent based on
the sample and mobile phase chemistries, making sure all
solutions/buffers are miscible and soluble. Remember: the
composition of the needle-wash solvent should be the most solubilizing
compatible solvent (your strongest diluent), selecting it is part of
the method development.
Generic Chromatographic
Condition Needle Wash Solvent
Buffered aqueous,
reversed-phase: 90% H 2 O, 10% MeOH
Non-aqueous,
reversed-phase:
100% MeOH
Normal phase Mobile phase/or
GPC: Mobile phase IPA or EtOH
Ion exchange:
H 2 O
Refresh
the syringe whenever you:
• Prime
the solvent management system
• Change
solvents
• See
bubbles in the syringe
• Start
using the 2795 Separations Module at the beginning of each day
• Washes
the inject port and the interior and exterior of the needle
degassed
purge solvent
Use the
Wash Needle function to:
• Ensure
that the needle and inject port are clean before the start of a new
method
flow
through the waste tube
Note: If
the 2795 Separations Module does not detect that a needle wash was
performed
since
the last time the 2795 Separations Module was powered on, it performs
an
automatic
needle wash before beginning any injections, either those specified in
a sample
set or
by using the Inject Samples direct function.
Note:
Waters recommends using a high percentage organic wash solvent that is
Attention:
Make sure the needle-wash solvent line is primed before you perform a
needle
8. Check for stability of
the pressure and enter the Run Samples (Quick set) in
Millennium/Empower to monitor the baseline.
Summary:
Steps in the Preparation of Alliance 2790/5
|
|
Screen |
|
|
A, B, C, D are mobile phase reservoirs and wash solvents
For "Needle
wash" solvent: use the most dissolving solvent For seal wash
solvent: use aqueous solvent. For Purge Solvent use
either diluent's or mobile phase's solvents composition
|
Check for solvents in:
Solvents in A, B, C, D; Purge
solvent Needle Wash Seal Wash |
|
|
Check
for pressure stability
|
Status:
Direct
Function - 2 |
Dry
Prime:
A, B, C, D. |
|
Build up the vacuum in
the degasser
|
Stop
Flow:
|
|
|
Verify full &
degassed vacuum chambers
|
Status:
Direct
Function - 2 |
Wet Prime again shortly.
|
|
Check for liquids in
yellow tubing outlet
|
Main:
|
Prime
Needle Wash
|
|
Check for liquids out
|
Main:
|
Prime
Seal Wash
|
|
1. Refresh
Syringe: wash the injector interior tubings with the diluent/mobile
phase solvents.
2. Wash Needle : wash the
injector interior with the strong solvent |
Status:
Direct
Function - 3 |
Refresh
syringe
|
|
Set the flow. It
will pass through the column.
|
Status
|
Set Flow
|