Guidance:

Using Tough Dirty and Complex Samples and/or  Mobile  Phases



Maintenance of HPLC/UPLC - Optical Detection

  • Sonication of all stainless-Steel filters in the system: 
    • 10% acetic acid in 45:45 water:Methanol followed by methanol then water (5 min each)
    • In worse cases use 10% nitric acid in water.

It is better to just replace the reservoir filters

  • Use "wash" vials: injection of full loop volume after each set of tough samples:
VARIOUS OPTIONS:

1. 20% acetic acid in water (for salts)

2. 5% NH4OH (for salts)

3. IPA (for lypophilic compounds)

4. Acetone (for polymeric residues)

5. DMSO:H2O (for bio-residues)

6. 1 mM EDTA (for metallic residues or chelating agents such as amino acids).

  • Purge injector in the end of every sample set using water/MeOH/MeCN mixtures according to the method
  • Wash the detector cell (using union or column's bypass) with the appropriate wash solution
  • Water (for salts)

  • Methanol or isopropanol (for oily residues)

  • 10% acetic acid in water (for salts)

  • 0.1 mM EDTA (for chelating agents)


Routine maintenance of HPLC-MS or ELSD

 

  1. Reservoir filters sonication of bottle A (H2O with FA) once a month
  2. In-line filter in the pump, sonication once a month
  3. Water for LC-MS should be clean of residues (check supplier)

 

Preventative measures:

 

  1. Use "wash" vials for injection of full loop volume after each set of tough samples:
OPTIONS:

1. 20% acetic acid in water (for salts)

2. 5% NH4OH (for salts)

3. IPA (for lypophilic compounds)

4. Acetone (for polymeric residues)

5. DMSO:H2O (for bio-residues)

6. 1 mM EDTA (for metallic residues or chelating agents such as catalysts).

  1. Purge injector in the end of every sample set:  (Masslynx: Use an inlet method with purge injector in the shutdown method)

Follow-up

 

  1. Inject 100ul 50:50 MeCN:H2O after every tough sample to wash contamination
  2. Inject blank (sample’s solvent) after every concentrated sample