GUIDANCE
for
Spectral
Analysis
using
Waters' Empower/Millennium
for Photo Diode Array (PDA)
Detection in HPLC
Definitions:
- The spectral contrast angle measures the shape difference
between two spectra.
It is the basis for all spectral comparisons in the software:
Bij =Spectrum
at each point in the peak
Ai = Spectrum at apex
s = normalization factor
0 <=  <=
1
0 <= 
<= 90 |
- The Purity angle is a measure of the spectral
heterogeniety of a
peak based on the comparison of spectra over all the peak, using the
spectral
contrast angle
- The non-ideal effects are quantified and provided as a
value of the Threshold
angle.
- When the peak is pure, the Purity angle is lower than the Threshold angle.
GENERAL
SPECTRAL ANALYSIS
(Not for validation of stability indicating chromatographic
method!).
Instrumental Considerations:
- Choose the spectral resolution preferred for the particular
compounds.
Preferably use 1.2 nm resolution for poor UV spectral shaped molecules.
When there are limitations in memory, select 2.4 or 3.6 nm, not higher.
- Choose the spectral range for the acquisition (nm).
Make sure that
it includes at least 5 x wavelength resolution values below the assay's
wavelength in the short wavelengths or 5x wavelength resolution values
above the assay's wavelength in the longer wavelengths.
- Check the mobile phase's UV cutoff wavelengths, and make
sure to work above
it.
- Select the appropriate acquisition rate (Number of spectra
per minute)
to make sure that there will be at least 15 points across the peaks.
Generally
1 spec/sec is sufficient, however, when fast chromatography is used 2
specs/sec
might be preferable (note the double storage memory needed).
Experimental Considerations
- If available, make sure to have a standard injected at
concentrations that
will give 0.5-1 AU in the MaxPlot chromatogram, make
sure
that the sample gives peaks at similar responses if available. The
standards
and the unknowns should be injected in the same session under the same
conditions exactly, to be able to evaluate the values of Threshold
angle and Purity
Angle.
- Make sure to include a blank run, to be able to subtract
the 3D data in
case of problematic results.
Software Considerations
- Create a method set that uses Maxplot derived channel and
apply it for
all the channels under study.
- Check carefully the entire chromatogram for quiet time
interval in the
chromatogram, minimum of 0.5 minutes long, to assign as the Noise
Interval.
- Create an appropriate PDA Processing Method that enables Peak
Purity,
set the appropriate wavelength range (cut the noisy low UV range or too
quiet high UV range). Set the Noise Interval and the criteria for
Threshold
Determination. The Auto-threshold will be applicable to most cases.
Results can be observed in the following places:
- Peaks tab at the bottom table in the main window of the Review.
- Results window, the top table, use the Peak Purity tab.
Also,
select
the
Purity Plot at the bottom, to observe the results
graphically.
- Using the Result view use the Preview
tool, select an existing
PDA report or create one with the report wizard, in which you can
select
the PDA Default Report.
SPECTRAL
ANALYSIS
(FOR VALIDATION OF STABILITY INDICATING METHOD)
Instrumental Considerations:
- Choose the spectral resolution preferred for the particular
compounds.
Preferably use 1.2 nm resolution for poor UV spectral shaped molecules.
When there are limitations in memory, select 2.4 or 3.6 nm, not higher.
- Choose the spectral range for the acquisition (nm). Make
sure that it includes
at least 5 x wavelength resolution values below the stability method's
wavelength in the low wavelengths and 5x resolution values above the
method
wavelength in the longer wavelengths.
- Check the solvent UV cutoff wavelenth and make sure to work
above it.
- Select the appropriate acquisition rate (Number of spectra
per minute)
to make sure that there will be at least 15 points across the peaks.
Generally
1 spec/sec is sufficient, however, when fast chromatography is used 2
specs/sec
might be preferable (note the double storage memory needed).
Experimental Considerations
- Make sure to have 5-6 replicates of the standard for the
Peak Purity analysis
and replicates of the standard at 25%, 50%, 75%, 100% 120% or others of
the assay's concentration for the spectral matching..
- The unknowns are stressed samples of the main component's
raw material
as well as the formulation. It is recommended to inject duplicates. The
samples stressed by heat, light, acid, base and oxidizing agent. Make
sure
to include their blank medium.
- The standards and the unknowns should be injected in the
same session with
the unknowns, under the same conditions exactly, to be able to evaluate
the values of Threshold angle from the Purity Angles
of the
standards.
- Make sure to include a blank run (diluent, mobile phase and
medium of the
stress) as well as the placebo, to be able to subtract the 3D data in
case
of problematic results.
Software Considerations
- Create a method set that uses Maxplot derived channel and
apply it for
all the channels under study.
- Apply the method set on all the runs, superimpose the
chromatograms in
the Results view window, and check carefully the entire
chromatogram
for quiet baseline interval in all the chromatograms, minimum of 0.5
minutes
long, to assign as the Noise Interval for all the runs.
Procedure:
Peak
Purity
Create an appropriate PDA Processing Method that
enables Peak
Purity, set the appropriate wavelength range (cut the noisy low UV
range or too quiet high UV range). Set the Noise Interval and the
criteria
for Threshold Determination (Auto-threshold or solvent+ noise). The
Auto-threshold
might not be applicable to all cases (because it makes assumptions
regarded
to the signal). For method validation in a regulated environment it is
recommended to use the manual sensitive Threshold angle determination:
1. Select the 6 standard replicates and process them, using
the method
set. Check the results for Purity Angle of all 6 runs. Select the
maximum
value of the Purity Angle and set it in the processing
method
under Solvent and Noise criteria for Threshold determination.
2. Process the 6 replicates again and verify that all
standards came
out pure (Purity Angle < Purity Threshold Angle).
3. Process the unknowns and use an appropriate PDA report to
report
their results. You can use the Purity Plots in the report. If
the Purity
angle is smaller than the Threshold angle, the peak is
pure.
Procedure:
Spectral
Matching
- Select one of each standard concentration (25%, 50%, 75%
and 100%), and
create a new spectral library. Make sure to indicate the standard
source
and level or concentration in its name for information.
- Check the library name in the Processing Method at the PDA
Spectral match
window. Select in this window the depth of search (1-3) and the degree
of match (in deg). Indicate if you are interested in RT presearch.
- Select the second channel in the duplicates from each
standard concentration
and process it using the method set with the updated processing method.
Check the Match angle of all 4 channels and select the maximum value
from
all match angles (3 results each channel), and use it in the Processing
Method as the Threshold angle for the Spectral Matching.
- Process the 4 Channels again, using the method set with the
updated Processing
Method. Verify that the results indicate that they all are identified
appropriately
(Match angle<Threshold angle).
- Process all the unknowns using the same method set. When
the Match angle
is smaller than the Threshold angle the match is OK.
Results can be
observed in the following places:
Peaks tab at the bottom table in the Main window of
the Review.
Results window, the top table, use the Library match tab.
Also,
select
the
Match Plot at the bottom, to observe the results
graphically.
Using the Result view use the Preview tool,
select an
existing PDA report or create one with the report wizard, in which you
can select the PDA Default Report.
BEWARE OF THE
FOLLOWING:
- The Noise's time interval must be peak-less
in all the chromatograms.
- The noise's time interval in gradients must
be near the studied
peak.
- Non linearity of Beer's law will result in
wrong results
and unusual behavior.
- Integration events can effect the results,
make sure to use
them appropriately.
See also:
Peak
Purity I ; Peak
Purity
II ; Peak
Purity
III ; Library
Matching .
Instrumental
Deviation
from
Beer's
Law
(Based on Waters' help in empower)