INTEGRATION EVENTS
Select a channel and look at it in the Review window. Open the appropriate Processing Method. The Default Processing method that is opened automatically, Untitled, in the Review window yields too many peaks and wrong integration in the example here:

Therefore there is a need for many changes in the Integration Event Table in the Processing method. Tile the two windows, the Main and the Processing Method's to be able to see the results of the change immidiately.

For an existing method, first change the Inhibit Integration ranges and then the global integration events: Threshold and Peak Width (T and PW). Use the icon (or menu Option) "Show Events" so you can change them graphically. Do not use the Threshold value to omit unwanted peaks, if you have too many peaks, make sure to use the appropriate T value to define all of them properly (Start and end are OK), then choose the "Minimum Height" (MH) option, by choosing the highest peak you want to omit, then clicking the MH button and adding 20% to the value, or use the "Minimum Area" option in the same manner:

Now that you have just the peaks of interest, you can start using the more advanced features of the integration events table in the Processing method window. The Table has four columns:
Time
Specifies the time at which the event type is to start. Integration event table rows are sorted automatically by the entered start times. To activate an event for an entire chromatogram, enter a value of 0.000 in the Time field and leave the Stop field blank. Valid entries: 0.000 to 655.000 minutes. Default: 0.000.
Type
Specifies the type of integration event to occur at the specified time. You can choose from 18 event types. The selected event begins at the time specified in the Time field and stops at the time specified in the Stop field.
Value
Specifies the value of an additional parameter required by some event types (for example, the Set Minimum Area event type requires the value of the minimum peak area in µV/sec).
Stop
Specifies the time at which the event is to stop. Events that use a
Stop time also allow a blank value, indicating activation of the event
until the end of the run. Certain events (such as Set Lift off, Set Touchdown)
do not use a Stop time. Valid entries: 0.000 to 655.000 minutes. Default:
Blank.
The Types column includes the following:
Integration events are classified as detection events and integration events.
Detection events affect single peak or fused-peak start and/or end points.
Detection events include:
Integration events include:
We will look one by one at the effect of the events as they appear in the list:
When a riding peak needs to be skimmed from the slope, there are two options, exponential or tangential skim:
Exponential Skim
For example, skim at 42 min. The value is a detection type of value - minimum slope to detect the riding peak.
Value 4. Value 10 (not effective)

Tangential Skim:
This event just connects the two starting and ending points of the riding peak
Force Drop line
For example: Force drop-line at 31.26 min to force fused peaks to split:
Originally Force Drop line

Remember that the place of the dropped line fits just the currently used chromatogram, it is very specific for a chromatogram. It is used frequently in GPC to define broad fused peaks, not so much in regular LC.
Force Baseline by Time & Force Baseline by Peak
Since we are dealing with a gradient chromatogram, the baseline might be drifting:

Projects a baseline based on either of the following:
· Event start and end times (Force Baseline by Time)
· Peak start and end points within this range (Force Baseline by Peak)
(Note: The forced baseline does not have to be horizontal.)
Force Base line by Peak at 50-59.4 min Force Base line by Time at 50-59.4 min

Force Peak
When adjacent peaks have to be integrated together and reported as a sum of areas:

Forward Horizontal by Peak (FHP) & Forward Horizontal by Time (FHP)
Baseline is forced to be horizontal from the first point in time forward (FHP) or from the first peak (FHP)
Forward Horizontal by time at 13.666-18.3 Forward Horizontal by peak at 13.666-18.3

Note that the "Force horizontal by Time" is very sensitive to the base line at the starting point of the event:
When the event started at 13.4 the picture was distorted due to the preceding peak:

Reverse Horizontal by Peak (FHP) & Reverse Horizontal by Time (FHP)
The same as Forward but backward.
Set Peak Width
The peak width is a global integration event that appears at the Review window and in the processing method as the first parameter to set. However, it might be changing throughout the chromatogram either gradually broadening at isocratic conditions or variable according to the focusing effects of the gradient. Therefore, it can be changed at various time segments in the chromatogram.
Global event PW = 30 Set peak width to 120 at 98 min

Set Liftoff & Set Touchdown
These two parameters come instead of the Threshold, which determines both peak liftoff and touchdown. If the global PT was not appropriate for a certain peak, it can be changed locally:
Threshold is at 300 Set Liftoff and Touchdown at 50 from 70 min.
Remember to set it back to 300 at 74 min

Valley to Valley
Sets the baseline at each valley start and end point in a fused peak (within the user-specified start and stop time window).
Note: Without this event, a common baseline is drawn for all fused peaks, with each peak separated by a drop line. When enabled, the Valley-to-Valley event reassigns the baseline at each peak start and end point. All peaks become baseline resolved and are labelled as BB in the Peaks table (Review Main window and Results window).

THE FINAL CHROMATOGRAM:
USING ALL THESE EVENTS:

And the Integration Events Table looks like that:

The Report will look like that:
Back to Waters' Customer Support in Israel
Back to Millennium advanced training
Medtechnica, Analytical Department
Efal St. 5 Petach Tikva Tel: 03-9254040 Fax: 03-9249977
Email
Home||
HPLC
Topics ||
LC-MS
TOPICS || HPLC
Course || Educational
|| Links
Courses
|| Personal
||
FTIR ||
Copyright ©2000 Dr. Shulamit
Levin